Hi,
It seems that automatic peak position hunting programs
like PowderX, EVA 5.1, Winplotr... are unable to
distinguish two
adjacent peaks when the maximum of
these 2 peaks are separated by only one
step, even if
the 2 peaks are clearly separated by eyes, seeing à
M-like
shape at the peak top like this :
|\/|
| |
_________| |__________
Comments ?
I have no experience with any of
these programs, but ran into similar
problems with commercial programs for
plotting and interpreting HPLC
chromatographs some years ago. There are
several possible problems:
1. The program may be smoothing the digital input
for its interpretation by
combining consecutive points.
2. The peak
detection may be set to require too many consecutive points to be
below the
maximum before a peak is detected.
3. The signal strength at peak onset may
be part of the calculations to
locate the peak, e.g., peak location is
midway between onset and end-of-peak,
and the peak end requires going back
to the signal strength at peak onset.
4. The peak may be determined by when
zero slope is detected, and the slope
calculation extends over more than two
adjacent points.
5. Was the baseline noisy? The program may be
automatically determining the
standard deviation of the signal from the
baseline noise, and the minimum
between the two peaks may be within that
statistical range of the maximums.
In any event, the problem could only
be answered in my case by a very
detailed examination of the peak-finding
code. If the programs you mention
have open code, that might be possible to
do. If the code is proprietary,
I'm not sure what the answer is.
Wilson DeCamp